User interface language: English | Español

Date November 2016 Marks available 2 Reference code 16N.3.HL.TZ0.19
Level Higher level Paper Paper 3 Time zone Time zone 0
Command term Show that Question number 19 Adapted from N/A

Question

The linear attenuation coefficient μ of a material is affected by the energy of the X-ray beam and by the density ρ of the material. The mass absorption coefficient is equal to \(\frac{\mu }{\rho }\) to take into account the density of the material.

The graph shows the variation of mass absorption coefficient with energy of the X-ray beam for both muscle and bone.

Show that the attenuation coefficient for bone of density 1800 kg m–3, for X-rays of 20 keV, is about 7 cm–1.

[2]
a.

The density of muscle is 1200 kg m–3. Calculate the ratio of intensities to compare, for a beam of 20 keV, the attenuation produced by 1 cm of bone and 1 cm of muscle.

[3]
b.

Suggest why more energetic beams of about 150 keV would be unsuitable for imaging a bone–muscle section of a body.

[1]
c.

Markscheme

reads value on graph at 20 keV as 4 «cmg–1»

«4 cmg–1 × 1800 kg m–3 × \(\frac{{1000}}{{1000000}}\) = » 7.2 «cm–1»

Ensure that the calculation has right POT conversion.

Answer must be to at least two significant figures.

a.

ALTERNATIVE 1
(finds intensity ratios for muscle and bone separately)
Watch for ECF

for muscle: obtains μ = 0.96 cm−1
Allow answers in the range of 0.90 to 1.02 cm–1.

\(\frac{I}{{{I_0}}}\) = eμx so for muscle 0.38

Allow answers in the range of 0.36 to 0.41.

Allow answers in dB. Muscle -4dB, Bone -30 or -31dB

for bone: \(\frac{I}{{{I_0}}}\) = 7.5 × 10−4 «if μ = 7.2 is used»

OR

9.1×10−4 «if μ=7 is used»

ALTERNATIVE 2
for muscle: obtains μ = 0.96 cm−1
Allow answers in the range of 0.90 to 1.02 cm–1.

\(\frac{{{I_{{\rm{MUSCLE}}}}}}{{{I_{{\rm{BONE}}}}}} = \frac{{{e^{ - 0.96}}}}{{{e^{ - 7.2}}}}\)

Frequently the POT will be incorrect for MP1. Allow ECF from MP1 to MP2.
Allow +/- 26 or 27dB
Award [2 max] if μ=960 as they will get \(\frac{{{I_{{\rm{MUSCLE}}}}}}{{{I_{{\rm{BONE}}}}}}\) = 0.

ratio is about 500 «513»
Allow range 395 to 546
If 7 used, ratio is about 420, if 7.2 is used, ratio is about 510
Allow answer IBONE/IMUSCLE from a range 0.0017 to 0.0026.

b.

similar absorption so poor contrast

c.

Examiners report

[N/A]
a.
[N/A]
b.
[N/A]
c.

Syllabus sections

Option C: Imaging » Option C: Imaging (Additional higher level option topics) » C.4 – Medical imaging (HL only)
Show 74 related questions

View options